Teledyne T802 Respiratory Product User Manual


 
Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
87
To access and navigate the VARS menu, use the following key sequence.
SETUP X.X 6) CLOCK_ADJUST=0 Sec/Day
PREV NEXT JUMP EDIT ENTR EXIT
SETUP X.X CLOCK_ADJUST=0 Sec/Day
+0 0 ENTR EXIT
Enter sign and number of
seconds per day the clock
gains (-) or loses(+)
SAMPLE RANGE=100.00 % O2=XXX.XX
<TST TST> CAL MSG SETUP
SETUP X.X 0) DAS_HOLD_OFF=15.0 Minutes
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X PRIMARY SETUP MENU
CFG ACAL DAS RNGE PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X ENTER PASSWORD:818
8 1 8 ENTR EXIT
Toggle these
buttons to enter
the correct
PASSWORD
In all cases:
EXIT discards the new
setting
ENTR accepts the
new setting
SETUP X.X 3) DYN_ZERO=OFF
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X 4) DYN_SPAN=OFF
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X DAS_HOLD_OFF=15.0 Minutes
1 5 .0 ENTR EXIT
Toggle these keys to set
the iDAS HOLDOFF time
period in minutes
(MAX = 20 minutes).
SETUP X.X 5) CONC_PRECISION=AUTO
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X CONC_PRECISION=AUTO
AUTO1234 ENTREXIT
Use these buttons to select
the precision of the O
2
concentration display
SETUP X.X 1) STABIL_GAS=O2
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X STABIL_GAS=NO
O2 CO2 ENTR EXIT
Use these buttons to select
which gas will be reported
by the STABIL test function
(CO2 is only available if
the optional CO
2
sensor is
installed)
SETUP X.X 2) TPC_ENABLE=ON
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X TPC_ENABLE:ON
ON ENTR EXIT
Toggle to turn on or turn
OFF temperature pressure
compensation.
07275B DCN6418