Teledyne T802 Respiratory Product User Manual


 
Calibration Procedures Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
178 07275A (DCN6005)
9.4.2. FLOW CALIBRATION
The flow calibration allows the user to adjust the values of the sample flow rates
as they are displayed on the front panel and reported through COM ports to match
the actual flow rate measured at the sample inlet. This does not change the
hardware measurement of the flow sensors, only the software-calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter
to the sample inlet. Once the flow meter is attached and is measuring actual gas
flow, press:
SETUP X.X PRIMARY SETUP MENU
CFG DAS ACAL RANG PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X ENTER PASSWORD:818
8 1 8 ENTR EXIT
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
Continue pressing NEXT until ...
DIAG FCAL ACTUAL FLOW: 120 CC/M
0120 ENTREXIT
Toggle to match the actual
flow as measured by the
external flow meter
EXIT discards the new
setting
ENTR accepts the
new setting
DIAG FLOW CALIBRATION
PREV NEXT ENTR EXIT
SAMPLE RANGE=100.00 % O2=XXX.XX
<TST TST> CAL SETUP
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
07275B DCN6418