Calibration Procedures Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
170 07275A (DCN6005)
9.1. BEFORE CALIBRATION
The calibration procedures in this section assume that the range mode, analog
range and units of measure have already been selected for the analyzer. If this
has not been done, please do so before continuing (see Section 5.4 for
instructions).
Note If any probl
ems occur while performing the following calibration
procedures, refer to Section 11 for troubleshooting tips.
9.1.1. REQUIRED EQUIPMENT, SUPPLIES, AND EXPENDABLES
Calibration of the T802 analyzer requires a certain amount of equipment and
supplies. These include, but are not limited to, the following:
Zero-air source.
Span gas source.
Gas lines - All Gas lines should be Stainless Steel, PTFE (Teflon), glass or
electroless nickel.
A recording device such as a strip-chart recorder and/or data logger (optional).
For electronic documentation, the internal data acquisition system can be
used.
9.1.2. CALIBRATION GASES
9.1.2.1. ZERO AIR
A gas that is similar in chemical composition to the earth’s atmosphere but
scrubbed of all components that might affect the analyzers readings. Teledyne
API recommends using pure N
2
when calibrating the zero point of your O
2
or
optional CO
2
sensor options.
CAUTION
GENERAL SAFETY HAZARD
DO NOT vent calibration gases into enclosed areas. Rapid release of pure N
2
gas
into an enclosed space can displace oxygen, and therefore represents an
asphyxiation hazard. This may happen with few warning symptoms.
9.1.2.2. SPAN GAS
Span gas is a gas specifically mixed to match the chemical composition of the
type of gas being measured at near full scale of the desired measurement range.
In this case, O
2
measurements made with the T802 analyzer, Teledyne API
recommends using 21% O
2
in N
2
when calibrating the span point of the O
2
sensor
option.
07275B DCN6418