Teledyne T802 Respiratory Product User Manual


 
Section II Operating Instructions Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
34
Channel A4 is special. It can be set by the user (see Section 5.9.6) to output any
one of the parameters accessible through the <TST TST> buttons of the unit’s
Sample display.
To access these signals attach a strip chart recorder and/or data-logger to the
appropriate analog output connections on the rear panel of the analyzer.
A
NALOG OUT
A1
A
2 A3 A4
+ - + - + - + -
Figure 3-7: Analog Output Connector
Table 3-5: Analog Output Pin-Outs
PIN ANALOG OUTPUT VOLTAGE SIGNAL CURRENT SIGNAL
1 V Out I Out +
2
A1
Ground I Out -
3 V Out I Out +
4
A2
Ground I Out -
5 V Out I Out +
6
A3
(Only used if CO
2
sensor is installed)
Ground I Out -
7 V Out NA
8
A4
Ground NA
3.3.1.4. CURRENT LOOP ANALOG OUTPUTS (OPTION 41) SETUP
This option adds isolated, voltage-to-current conversion circuitry to the
analyzer’s analog outputs. If your analyzer had this option installed at the
factory, there are no further connections to be made. Otherwise, it can be installed
as a retrofit for each of the analog outputs of the analyzer . This option converts
the DC voltage analog output to a current signal with 0-20 mA output current.
The outputs can be scaled to any set of limits within that 0-20 mA range.
However, most current loop applications call for either 2-20 mA or 4-20 mA
range. All current loop outputs have a +5% over-range. Ranges with the lower
limit set to more than 1 mA (e.g., 2-20 or 4-20 mA) also have a -5% under-range.
Figure 3-8 provides installation instructions and illustrates a sample combination
of one current output a
nd two voltage outputs configuration. The section
following
this provides instructions for converting current loop analog outputs to
standard 0-to-5 VDC outputs. Information on calibrating or adjusting these
outputs can be found in Section 5.9.3.7.
07275B DCN6418