Teledyne T802 Respiratory Product User Manual


 
Basic Operation Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
64
Figure 4-1: Front Panel Touchscreen and Display
The mode field of the front panel display indicates to the user which operating
mode the unit is currently running.
In addition to SAMPLE and SETUP, other modes the analyzer can be operated
in are:
Table 4-1: Analyzer Operating Modes
MODE EXPLANATION
SAMPLE Sampling normally, flashing text indicates adaptive filter is on.
SAMPLE A
Indicates that unit is in Sample Mode while AUTOCAL feature is active.
O2 M-P CAL This is the basic calibration mode of the instrument and is activated by pressing the CAL key.
SETUP [X.X]
SETUP mode is being used to configure the analyzer. The gas measurement will continue during this
process. The revision of the T802 firmware being run will appear after the word “SETUP
CAL O2 Z[type]
2 & 3
Unit is performing O
2
ZERO calibration procedure.
CAL O2 S[type]
2 & 3
Unit is performing O
2
SPAN calibration procedure
CAL CO2 Z[type]
2 & 3
Unit is performing CO
2
ZERO calibration procedure (when the optional CO
2
sensor is installed).
CAL CO2 S[type]
2 & 3
Unit is performing CO
2
SPAN calibration procedure (when the optional CO
2
sensor is installed).
DIAG Mode One of the analyzer’s diagnostic modes is active (Section 5.9).
[type:]
2
M: initiated manually by the user via the front panel touchscreen.
3
R: initiated remotely through the COM ports or digital control inputs.