Teledyne T802 Respiratory Product User Manual


 
Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
59
3.4.3.3. SET O
2
SPAN GAS CONCENTRATION
Set the expected O
2
span gas concentration. This should be 80% of concentration range for which the
analyzer’s analog output range is set.
O2 M-P CAL O2 SPAN CONC:20.95 %
0 2 0 .9 5 ENTR EXIT
SAMPLE GAS TO CAL:O2
O2 CO2 ENTR EXIT
SAMPLE RANGE=100.00 % O2=XXX.XX
<TST TST> CAL CALZ CALS SETUP
O2 M-P CAL RANGE=100.00 % O2=XXX.XX
<TST TST> ZERO SPAN CONC EXIT
SAMPLE RANGE TO CAL:RNG1
RNG1 RNG2 ENTR EXIT
The O
2
span concentration value is
automatically default to
20.95 %.
If this is not the the concentration of
the span gas being used, toggle
these buttons to set the correct
concentration of the O
2
calibration
gas.
EXIT ignores the new
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the
CONCENTRATION
MENU.
Only appears if either
the optional CO
2
sensor
is installed.
Only appears if the
analyzer is set for
DUAL range mode
07275B DCN6418