Teledyne T802 Respiratory Product User Manual


 
Section II Operating Instructions Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
48
ATTENTION
COULD DAMAGE INSTRUMENT AND VOID WARRANTY
Remove dust plugs from rear panel exhaust and supply line fittings
before powering on/operating instrument. These plugs should be
kept for reuse in the event of future storage or shipping to prevent
debris from entering the pneumatics.
CAUTION
GENERAL SAFETY HAZARD
The exhaust from the analyzer’s internal or customer supplied external
pump MUST be vented outside the immediate area or shelter surrounding
the instrument.
Calibrated N
2
100%
Concentration
Calibrated O
2
at 20.95% Span
Concentration
VENT
Figure 3-16: Pneumatic Connections–Basic Configuration–Using Bottled Span Gas
3.3.2.5. SAMPLE GAS SOURCE
Attach a sample inlet line to the SAMPLE inlet port. The SAMPLE input line
should not be more than two (2) meters long.
Maximum pressure of any gas at the SAMPLE inlet should not exceed 1.5 in-
Hg above ambient pressure and ideally should equal ambient atmospheric
pressure.
In applications where the sample gas is received from a pressurized
manifold, a vent must be placed on the sample gas before it enters the
analyzer. Please refer to Figure 3-16.
07275B DCN6418