Analyzer Operation Manual Calibration Procedures
07275A (DCN6005) 171
Cylinders of calibrated O
2
gas traceable to NIST-Standard Reference Material
specifications (also referred to as SRMs or EPA protocol calibration gases) are
commercially available.
Table 9-1: NISTSRM's Available for Traceability of O
2
Calibration Gases
NIST-SRM Type Nominal Concentration
2657a
O
2
in N
2
2%
2658a
O
2
in N
2
10 %
2659a
O
2
in N
2
21%
2619a
1
CO
2
in N
2
0.5%
2620a
1
CO
2
in N
2
1%
2622a
1
CO
2
in N
2
2%
2624a
1
CO
2
in N
2
3%
2744b
1
CO
2
in N
2
7%
2745
1
CO
2
in N
2
16%
1
Used to calibrate optional CO
2
sensor.
Note
For span point calibration it is generally a good idea to use 80% of the reporting
range for that channel. For instance, if the reporting range of the instrument is
set for 5%, the proper span gas would be 4%
.
9.1.3. DATA RECORDING DEVICES
A strip chart recorder, data acquisition system or digital data acquisition system
should be used to record data from the serial or analog outputs of the T802.