Teledyne T802 Respiratory Product User Manual


 
Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
73
5.4.3.1. RNGE MODE SNGL: CONFIGURING THE T802 ANALYZER FOR SINGLE
RANGE MODE
Single Range Mode (SNGL) is the default reporting range mode for the analyzer.
When the single range mode is selected (SNGL), all analog O
2
concentration
outputs (A1 and A2) are slaved together and set to the same reporting range
limits (e.g. 0- 22.00 %). The span limit of this reporting range can be set to any
value within the physical range of the analyzer.
Although both outputs share the same concentration reporting range, the
electronic signal ranges of the analog outputs may still be configured for different
values (e.g. 0-5 VDC, 0-10 VDC, etc; see Section 5.9.3.1)
To select SNGL range m
ode and to set the upper limit of the range, press:
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
07275B DCN6418